WEF is pleased to announce the creation of free LinkedIn groups associated with its technical discussion forum topics. Through these LinkedIn groups, members will be able to view and participate in discussions, as well as communicate with one another via direct messages. Share your experiences and knowledge, ask questions and respond to other discussions as frequently as you like!

You must have a LinkedIn profile to join a LinkedIn group; you can create your profile here. If you already have a LinkedIn profile, all you need to do is submit a request to join one or all of the following groups:

Water Environment Federation (Main) | Biosolids | Collection Systems | Nutrients | Stormwater 
Utility Management | Water Reuse | Water for Jobs | Watershed Management | Laboratory Practices 

RSS Feed Print
BOD standard duplicates discrepancies
Tom Hons
Posted: Sunday, May 16, 2010 5:55 PM
Joined: 5/16/2010
Posts: 6


Hi,  For quite awhile, the results from the BOD standard duplicates (with identical seed and quantity of standard) have shown a fairly consistent pattern that the first bottle will show about 10-30% higher DO depletion than the 2nd.   Thoughts?


Perry Brake
Posted: Sunday, May 16, 2010 11:37 PM
Joined: 12/16/2009
Posts: 69


You don't give much information upon which to make a guess.  If you can provide the following info, it might help.

 

1.  What is the magnitude of the difference?

 

2.  Do you do duplicate blanks, and if so, is there a similar difference between the two blank results?

 

3.  In what sequence do you set up and read the bottles in a batch?

 

4.  What is the relationship between the first and second GGA bottles with regard to location in the incubator?


Tom Hons
Posted: Monday, May 17, 2010 2:05 PM
Joined: 5/16/2010
Posts: 6


To answer the questions, 1) the first bottle will often have about a 0.3 to 0.8 mg/l higher depletion.  

2) I don't incubate one of the 2 blanks. I just use it to get an initial DO.

3) Order is  blanks, standards, seed , final,  raw, others

4) All bottles are in a 12 bottle carrier. the standards are the 2nd and 3rd in back corner after the blank. The incubator is the size of a 1/2 size refrig. (like undercabinet)


Perry Brake
Posted: Monday, May 17, 2010 4:42 PM
Joined: 12/16/2009
Posts: 69


Hmmm...

 

I asked about the sequence of bottles in the batch to see if perhaps you were doing the seed control just before the GGA, which could lead to carryover...but that's not the case.

 

You might try switching the order of the GGA bottles in the 12-bottle carrier.  One of the two bottles might be a tad warmer than the other because of proximity to a source of heat in the incubator (a converted fridge?).  If the second bottle then consistently has higher results than the first, the permanent solution might be to put a small fan in the incubator to circulate the air, to place the entire carrier in a different location in the incubator (if possible), or to replace the incubator.

 

One other thing I should have asked is whether you use a DO meter/probe, or Winkler titration to measure DO.  Unless you are doing Winkler's, there is no need to take an initial DO on one blank bottle and incubate the other one.  If you are using a meter/probe, you might try measuring both initial and 5-day DOs on both bottles and see if you experience one bottle always being a little higher than the other.  Alternatively, you could do a few effluent samples in duplicate.  In the long run, however, you don't get much useful information from doing blanks in duplicate, and the same can be said for doing effluent in duplicate if your precision is good (as indicated by a low standard deviation on GGA results.

 

There might be some other cause, but I sure can't think of what it might be.

 

 


Dean Falkner
Posted: Sunday, August 28, 2011 8:26 AM
Joined: 6/21/2011
Posts: 11


Hi Tom:

 

Are the GGA samples the first bottles filled?  I've seen cases where the dilution water tubing gets a growth on it.  The initial samples got contaminated as some of this growth was rubbed off during the first bottle filling.  Periodic bleaching eliminated the issue.

 

Dean