WEF is pleased to announce the creation of free LinkedIn groups associated with its technical discussion forum topics. Through these LinkedIn groups, members will be able to view and participate in discussions, as well as communicate with one another via direct messages. Share your experiences and knowledge, ask questions and respond to other discussions as frequently as you like!

You must have a LinkedIn profile to join a LinkedIn group; you can create your profile here. If you already have a LinkedIn profile, all you need to do is submit a request to join one or all of the following groups:

Water Environment Federation (Main) | Biosolids | Collection Systems | Nutrients | Stormwater 
Utility Management | Water Reuse | Water for Jobs | Watershed Management | Laboratory Practices 

RSS Feed Print
MDL Quandary
Keith Chapman
Posted: Wednesday, April 28, 2010 12:30 PM
Joined: 10/2/2009
Posts: 35


Which of these two MDL's for Beryllium by ICPOES is "most correct"?  We are following the EPA (40CFR136 Appendix 2?) procedure.

 

1.  Using a concentration of 0.2 ppb gives an MDL = 0.04 ppb

2.  Using a concentration of 0.1 ppb gives an MDL = 0.08 ppb

 

Both MDLs meet the criteria that Tested Concentration and MDL must be within a ratio of from 1:1 to 1:10.

 

Which MDL is "best"?


Perry Brake
Posted: Wednesday, April 28, 2010 1:43 PM
Joined: 12/16/2009
Posts: 69


Appendix "2" is almost right, Keith!  It's Appendix B.

 

http://http://ecfr.gpoaccess.gov/cgi/t/text/text-idx?c=ecfr&sid=64839efb3f45e2c705aef2259e906baf&rgn=div9&view=text&node=40:22.0.1.1.1.0.1.7.2&idno=40 

 

You don't mention how you came up with the 0.2 and 0.1 ppb concentrations to use in the MDL study, but I guess one can assume it was by one of the following three procedures specified in Appendix B...

 

1. Make an estimate of the detection limit using one of the following:

(a) The concentration value that corresponds to an instrument signal/noise in the range of 2.5 to 5.

(b) The concentration equivalent of three times the standard deviation of replicate instrumental measurements of the analyte in reagent water.

(c) That region of the standard curve where there is a significant change in sensitivity, i.e. , a break in the slope of the standard curve.

 

In sub-paragraph 2 immediately following the quoted material above, App B says to prepare a solution that is 1 to 5 times, the estimated MDL.  Without knowing what your estimated MDL was, I don't think anybody can tell which of your values...if either...is correct or "best."

 

Assuming that both your 0.1 and 0.2 ppb solutions meet the "estimated MDL" criteria, I would conclude that the 0.2/0.04 situation is "best".  App B cautions that choosing an estimated MDL that is too low might lead to an inflated MDL, and one could use that as justification for not choosing the 0.1 ppb situation as being "best".

 

 

 

 

 

 

 

 


Charles Lytle
Posted: Friday, May 7, 2010 8:11 AM
Joined: 10/5/2009
Posts: 49


Keith's data actually are two points on a "continuum" that's highest at the smallest allowed spike concentration and should go to zero as the spike concentration gets larger and larger.  The careful use of significant figures should make the disappearance occur fairly quickly.  (Think about this before you climb all over my bones.)

 

I can't remember how the EPA arrived at its range of acceptable spike/MDL ratio, but it has to do with coming to some sort of happy medium.  I'd be fun to do a whole bunch of studies and actually plot the calculated MDLs as the spike concentration varies.  Maybe somebody already did this??

 

Chuck Lytle

 

Added later:  I forgot to address Keith's question.  I agree with Perry.


Charles Lytle
Posted: Monday, May 10, 2010 3:31 PM
Joined: 10/5/2009
Posts: 49


To further my point, in looking through a binder of articles on detection limits, I found the following, taken from the September 2006 Application Notebook that comes as an insert with Spectroscopy magazine...

 

Following the procedure at 40 CFR 136 for nickel by ICP, the authors got the following two results:

 

IDL = 0.0025 ppm

MDL = 0.0018 ppm

 

One would expect that, as you add confounding factors such as reagents, digestion procedures, etc., precision should fall off.  Not the case, here.

As Mark Twain observed, "there are lies, damned lies, and statistics."  happy

 

Chuck Lytle


hubblijr
Posted: Tuesday, May 11, 2010 3:25 PM
Joined: 5/11/2010
Posts: 3


We run into this with ICP/MS MDL determinations all the time.  The analyses are so easy and fast to run, we run 10 standards for all metals and will usually see two to three different but acceptable MDL's for each metal.  But as pointed out, at the MDL you have either zero or one sig fig (depending on if you are a chemist or statistician) so be careful of meanless sensitivity.

 

The problem with the MDL (and that has led to lawsuits and the recently closed Fed. Advise. Committee on Hg MDL) is the assumption that noise (as sigma, or %RSD) is constant over the range from zero to 5 times the MDL; it's not.  Only the ASTM D19 method for "Detection Estimate" addresses this issue (that I know of) by modeling uncertainty with concentration at low concentrations allowing one to pick the "exact" point when the det. limit level is reached (usually 33% RSD for 3 sigma uncertainty).

 

As appendix B suggests, this mulitple MDL problem can be addressed by iteratively converging on the single correct point (output from one study is input to the next). 

 

Unfortunately, there is a huge disconnect between the practical use of data and compliance with regulatory limits at levels pushing detection.  Data even at 3 times the MDL has much more uncertainty involved that is generally recognized.