Joined: 4/4/2012 Posts: 1
|
Does anyone have a reference for calculating matrix spike recoveries when the analyte in the unknown sample is at the method detection limit? All the calculations I have seen instruct you to subtract the 'measured concentration of analyte before spiking' from the 'measured concentration of analyte after spiking'. What happens when the analyte is recovered below your PQL or, more important, below your MDL? Do you subtract the actual concentration, even if it is a slightly negative number (which does sometimes happen on the ICP-AES) or do you subtract the MDL concentration or do you not subtract anything at all?
|
Joined: 10/5/2009 Posts: 45
|
Ruby,
Here's a quote from the EPA Inorganics CLP Statement of Work. It's in Section 12.6.7 of Exhibit C:
"When the sample concentration is less than the Method Detection Limit (MDL)...use zero...for purposes of calculating percent recovery."
There may be other references, but this particular one happened to be on my desk.
Chuck Lytle
City of Portland, OR
|
Joined: 1/18/2012 Posts: 6
|
We use zero in this case as well.
|