WEF is pleased to announce the creation of free LinkedIn groups associated with its technical discussion forum topics. Through these LinkedIn groups, members will be able to view and participate in discussions, as well as communicate with one another via direct messages. Share your experiences and knowledge, ask questions and respond to other discussions as frequently as you like!

You must have a LinkedIn profile to join a LinkedIn group; you can create your profile here. If you already have a LinkedIn profile, all you need to do is submit a request to join one or all of the following groups:

Water Environment Federation (Main) | Biosolids | Collection Systems | Nutrients | Stormwater 
Utility Management | Water Reuse | Water for Jobs | Watershed Management | Laboratory Practices 

RSS Feed Print
Orion 4 Star Ammonia ISE issue
Amy Vail
Posted: Monday, July 19, 2010 10:33 AM
Joined: 7/19/2010
Posts: 1


I am trying to calibrate a new Orion probe #9512 using a new Orion 4-Star meter.  Every time I try the standards read almost opposite of what they should (1 ppm reads around 15 ppm, 10 ppm reads around 5 ppm, and 100 ppm reads around 0.8 ppm).  I've remade the standards, used an older probe of the same model, etc.  Any suggestions?


Michele Melendez
Posted: Tuesday, July 20, 2010 7:46 AM
Joined: 10/2/2009
Posts: 1


I haven't performed ammonia analysis by ISE in awhile, but I'll try to help

 

We don't use the Orion 4 star meter, as we have an older Orion model.  But, if I recall correctly, we had the same sort of problem when we attempted to calibrate it for the first time.  Are you following through the entire calibration procedure or are you stopping as soon as you get wacky readings?  If you are stopping, try following through on the entire calibration procedure even though your readings are strange.  If you obtain an acceptable slope, then analyse a check standard and see if it is reading correctly.  You may be suprised to find that it is. 


James Royer
Posted: Wednesday, July 21, 2010 1:05 PM
Joined: 9/21/2009
Posts: 98


The ammonia gas sensing electrode has a negative slope so until you get a negative slope in the machine the results will look bad. After calibration the results should be good. Some old machines we had to tell it that the slope was approximately a certain value as it could only make small adjustments.


Anonymous
Posted: Monday, January 3, 2011 3:42 PM

HELP I AM NEW LAB SUPERVISOR AND AM TRYING TO UP DATE METTERS. I HAVE LOOKED AT ORION,HATCH AND YSI METTERS WHICH BRAND IS BETTER IN A LAB SITUATION


Rashmi Patel
Posted: Thursday, July 28, 2011 10:20 AM
Joined: 4/1/2010
Posts: 3


Amy,

 

Perhaps you have already addressed the situation with the NH3 probe by now.  However, I would like to add that it is a good idea to run the probe function test if it is determnined that neither the standards nor the meter (at the extreme) are the cause of the irratic or erroneous readings.

 

Happy testing,

 

Manlab