<?xml version="1.0" encoding="UTF-8"?>
<rss version="2.0">
 <!-- Generated by Ektron CMS400.NET -->
 <channel>
  <title>WEF Discussion Forums : Laboratory Management and Technical Issues : GGA always near 150 mg/L.</title>
  <description></description>
 <item>
  <title></title>
  <description><![CDATA[<root><p><font size="2"><font face="Verdana">Hi!</font></font></p>
<p><font size="2"><font face="Verdana"> </font></font></p>
<p><font size="2"><font face="Verdana">I've done everything and I can't get to the 198 mg/L value for my GGA. Here is a list of what I've done:</font></font></p>
<p><font size="2"><font face="Verdana"> </font></font></p>
<p><font size="2"><font face="Verdana">I'm using polyseed.</font></font></p>
<p><font size="2"><font face="Verdana">Hydrate/ aerate seed for 1h, 2h, overnight. Overnight one add few mL of GGA as food. For every test, the following volumes of seed have been tested, 2, 3, 4 and 5 mL. For some test I've tried even 10 mL. </font></font></p>
<p><font size="2"><font face="Verdana"> </font></font></p>
<p><font size="2"><font face="Verdana">-Different lot number of seed, same results</font></font></p>
<p><font size="2"><font face="Verdana">-Try hydratation of seed with 35 C of dilution water.</font></font></p>
<p><font size="2"><font face="Verdana">-Try 1, 2 or 3 capsule of polyseed.</font></font></p>
<p><font size="2"><font face="Verdana">-Try 2% and 1% GGA dilution same results</font></font></p>
<p><font size="2"><font face="Verdana">-Used a different source water, purchased Steam distilled water from Nursery. This water had fluoride and minerals added. From this test got values a bit higher for the 4mL of seed, 155, 158, 157 mg/L and 5 mL of seed, 157, 157 and 158mg/L for 2% GGA solution. Still not much of an improvement.</font></font></p>
<p><font size="2"><font face="Verdana"> </font></font></p>
<p><font size="2"><font face="Verdana">Blanks are below 0.2mg/L, and the seed depletion/mL are between 0.21 to 0.35 mg/L per mL of seed. </font></font></p>
<p><font size="2"><font face="Verdana"> </font></font></p>
<p><font size="2"><font face="Verdana">I've ask to colleagues to analyze source water for heavy metals, all were below MDL (Cu lower then 2 ppb), and for TOC lower then 0.2ppm. </font></font></p>
<p><font size="2"><font face="Verdana"> </font></font></p>
<p><font size="2"><font face="Verdana">Could the nutrient solutions played a role in affecting the GGA results? Any suggestions are welcome...</font></font></p>
<p> </p></root>
<br />]]></description>
  <pubDate>Wed, 20 Apr 2011 19:00:27 GMT</pubDate>
 </item>
 <item>
  <title></title>
  <description><![CDATA[<p>Your seed depletion of 0.21 - 0.35 is very low. The seed depletion should be between 0.6 - 1.0. I use a seed from NCL with very good results. You also might try using settled primary effluent for seed.</p>
<br />]]></description>
  <pubDate>Thu, 21 Apr 2011 13:25:35 GMT</pubDate>
 </item>
 <item>
  <title></title>
  <description><![CDATA[<root><p>Polyseed has no nitrifying bacteria by design, so when you think you are doing a BOD, you are actually doing a <strong>C</strong>BOD since the nitrogen-containing materials in the sample (the glutamic acid in the case of GGA) are not oxidized to nitrate.  So you would <em>expect </em>to get low GGA results with Polyseed.</p>
<p> </p>
<p>The commercial seed sold by NCL (mentioned in the previous post) is made by BioSystems which, according to the manufacturer, also has no nitrifiers.  However, my experience in evaluating several labs using BioSystems seed is that it usually performs better than Polyseed.</p>
<p> </p>
<p>If your lab is at a wastewater treatment plant, the <em>best </em>seed is close at hand, and free.  Primary effluent (such as runoff from a clarifier) should have a good population of hungry, healthy bacteria.  The supernatant liquid from settled sludge is sometimes a good source of bacteria.  Raw influent sometimes works, but is often very inconsistent because of natural and human activities that influence the bacteria.  If your wastestream includes industrial waste, watch for the possibility of toxicity that might weaken or kill the bacteria.</p>
<p> </p>
<p>My best advice to you, if you are at a WWTP, is to try taking your seed from your own wastestream and use a commercial seed as a last resort.  If you are at a commerical lab, try using a seed from a local WWTP that is doing well on its own GGAs.  If that works, you may want to use that seed as a permanent fix...I know may commericial labs that have had success doing that, keeping the seed sample in their incubator until a control chart shows it is starting to weaken.</p>
<p> </p>
<p>Good luck!<br /></p></root>
<br />]]></description>
  <pubDate>Thu, 21 Apr 2011 14:38:01 GMT</pubDate>
 </item>
 <item>
  <title></title>
  <description><![CDATA[<root><p>We use the Biosystems seed material and it seems to work with more consistency than the polyseed. It is a higher demand seed material, thus more media is left with the bacteria, and they seem to recover from dehydration better. You can also use settled fresh sewage if there are not an excess of nitriers. We get near the 198 value for both BOD and CBOD.</p>
<p> </p>
<p>I would surely get a new lot of GGA standard, maybe from another manufacturer, to ensure the standard is not in error. You can try new nutrient solutions, or nutrients from Hach already mixed, for comparison. The cost is not much. Find a lab that does not have a problem with BOD standards and have them use some of your source water to see if that is the cause.</p>
<p> </p>
<p>Using seed material that contains nitrifiers is not the answer as ammonia from the sample and ammonia from the dilution water will both contribute to the result. This is why the BOD test has such notorious reputation as a bad procedure. Different dilutions yeild big diferences making comparison among labs difficult.</p></root>
<br />]]></description>
  <pubDate>Fri, 22 Apr 2011 14:07:29 GMT</pubDate>
 </item>
 <item>
  <title></title>
  <description><![CDATA[<root><p>Are you making your GGA yourself?  If so, are you desiccating the glucose and glutamic acid prior to weighing out?  If you don't desiccate to remove excess moisture you will actually be making a standard at a lower concentration.</p>
<p>Your SCF does seem very low to me as well.</p>
<p>Joe</p></root>
<br />]]></description>
  <pubDate>Fri, 22 Apr 2011 14:40:11 GMT</pubDate>
 </item>
 <item>
  <title></title>
  <description><![CDATA[<root><p>Thanks everyone for the suggestions. We will try the seed you requested Perry.</p>
<p> </p>
<p>Joe, we are drying the GGA prior to making the solution at 103C for 1h. </p>
<p> </p>
<p>I will come back as soon as I have more results to share with you.</p>
<p> </p>
<p>Cheers</p>
<p> </p>
<p>Josée</p></root>
<br />]]></description>
  <pubDate>Thu, 28 Apr 2011 11:37:06 GMT</pubDate>
 </item>
 <item>
  <title></title>
  <description><![CDATA[<root><p>I've tested the B.O.D.Seed and the GGA is now near 160mg/L. An improvement, I guess. Other then that we could not get the 198 mg/L magic number for the GGA. </p>
<p> </p>
<p>Josée</p></root>
<br />]]></description>
  <pubDate>Tue, 31 May 2011 17:57:11 GMT</pubDate>
 </item>
 <item>
  <title></title>
  <description><![CDATA[<p>Josée, you haven't said whether you are doing BOD or <strong>C</strong>BOD.  I think we have all assumed you are doing BOD, but perhaps you are doing CBOD.  If so, a 160 mg/L average is not bad.<br /></p>
<br />]]></description>
  <pubDate>Wed, 01 Jun 2011 16:12:47 GMT</pubDate>
 </item>
 <item>
  <title></title>
  <description><![CDATA[<root><p>Perry, we are doing BOD but I guess since we are using B.O.D. Seed (Or polyseed) and they have no nitrifing bacterias, it like doing CBOD for the GGA. Also since we d'ont add nitrification inhibitor to our sample we have to report BOD. In addition we analysed our PT samples and we were not even near the CBOD values.</p>
<p>our BOD values: 26, 59, 93, 117 mg/L done with polyseed at the time</p>
<p>BOD values expected: 38, 76, 125, 148 mg/L</p>
<p>expected CBOD values: 34, 70, 115, 136 mg/L</p>
<p>We are still off for the CBOD.</p></root>
<br />]]></description>
  <pubDate>Mon, 06 Jun 2011 12:48:21 GMT</pubDate>
 </item>
 <item>
  <title></title>
  <description><![CDATA[This is exactly what I was lokonig for. Thanks for writing!
<br />]]></description>
  <pubDate>Mon, 27 Jun 2011 06:54:29 GMT</pubDate>
 </item>
 </channel>
</rss>
